Thread: How to change E3641a into normal mode from limit mode in remote status


Permlink Replies: 2 - Pages: 1 - Last Post: Mar 20, 2013 6:19 AM Last Post By: kt
selionz

Posts: 1
Registered: 06/28/12
How to change E3641a into normal mode from limit mode in remote status
Posted: Jun 28, 2012 9:54 AM
Click to report abuse...   Click to reply to this thread Reply
Attachment E3641.JPG (94.7 KB)
I am writing a ATE SW to control E3641a power supply with labview.
I setup its output at 12V. When it is running the display always shows voltage(12V) and limit current(0.5A) but my customer want it shows real current instead of current limit.
If it is at local status, to change between “limit mode” and “normal mode” just press button of “Display Limit” is fine. The problem is I cannot find any command it works as the function of “Display Limit” button in remote status.
It would be appreciated if anyone can help me out.

Selion
Maze

Posts: 1
Registered: 02/25/13
Re: How to change E3641a into normal mode from limit mode in remote status
Posted: Feb 25, 2013 2:12 PM   in response to: selionz in response to: selionz
Click to report abuse...   Click to reply to this thread Reply
Hi Selion,

i have the same problem..

Have you managed to solve the problem?

Cumps
kt

Posts: 5
Registered: 03/20/13
Re: How to change E3641a into normal mode from limit mode in remote status
Posted: Mar 20, 2013 4:29 AM   in response to: Maze in response to: Maze
Click to report abuse...   Click to reply to this thread Reply
There is no available SCPI command for "display limit". There are 2 ways to clear this "limit". 1. Manually press Local Key and clear the "limit". 2. By sending "*RST" command, it will reset back to "default" setup.Hope this help.

Edited by: kt on Mar 20, 2013 6:19 AM